Housekeeping Standard: Beryllium work areas shall be kept as clean as the nature of the work allows. However, during nonoperational periods, residual contamination shall be verified to be no greater than 3 µg Be/100 cm2 (this is the housekeeping standard). If contamination is found to be greater than this limit, then the area in question shall be recleaned and reevaluated to determine compliance with the housekeeping standard. See Section 3.8 for actions necessary for compliance with this standard.
|
Release Standard for Contaminated Materials: Material contaminated with beryllium may be released to non-beryllium work areas, provided that the material is first cleaned, the beryllium level is verified to be less than 0.2 µg Be/100 cm2, and the material is labeled as required in Section 3.5.8 (Figure 3). The term "contaminated" means having removable surface contamination that is in excess of whichever of the following is higher:
|
* 0.2 µg Be/100 cm2.
|
* The concentration of beryllium in the soil at the point of release (when there is visible accumulation of debris).
|
Environmental Standards: The U.S. Environmental Protection Agency (EPA) has established the following limits:
|
* Maximum permissible air concentration level, which is 0.01 µg Be/m3 averaged over a 30-day period (in 40 CFR 61.32, Emission Standard).
|
* Maximum contaminant level for beryllium in drinking water, which is 4 µg Be/L (in 40 CFR 141.62).
|
However, the EPA has not yet set water quality criteria for beryllium in ambient water (40 CFR 131.38). The regulations specify that individual National Pollutant Discharge Elimination System permits shall address beryllium under the EPA's narrative criteria for toxics. In addition, LLNL has established reuse criteria for soils generated and reused at LLNL (see Document 33.3, "Management of Soils and Debris," in the ES
|
3.4 Engineering Controls
|
This section specifies the engineering controls applicable to the handling of beryllium and its compounds. Maintenance of systems used to control beryllium are subject to all the controls established in this document to minimize worker exposure to beryllium and to minimize opportunities for exposure.
|
Section 3.7, Facilities and Equipment, provides additional requirements for facility design and layout.
|
3.4.1 Exhaust Ventilation
|
All repetitive indoor operations that may generate airborne beryllium equaling or exceeding the AL shall be enclosed and shall be equipped with local exhaust ventilation. Exhaust systems used to control beryllium work shall comply with Document 12.2 and 12.4. The ES
|
3.4.2 HEPA Filtration
|
Required HEPA filter exhaust systems shall comply with Document 12.5 which implements "HEPA Filter and In-place Leak Testing Standard," (UCRL-AR-133354, Rev. 2). Required HEPA filter systems shall be visually inspected semiannually and be evaluated for performance annually.
|